Alaris Air in Line Setting
We have been getting some reports from nursing of large air bubbles making it past the air sensor in our Alaris pumps without an alarm. Our current air in line setting for our adult profile is 250mcL and has been this was since we started using the Alaris pumps in 2017. The factory setting from Alaris is 125mcL but they don't have a best practice recommendation on what to set this at. If your health system uses Alaris, what is your air in line setting for your adult profile?
Thanks!