NICU 3-mL Doses - Syringe Size and Pump Flow Rate Limitation

PLEASE NOTE:   Posts made to this forum should not be considered as the expressed opinions of, nor should be considered endorsed by, the Medication Safety Officer’s Society (MSOS) or the Institute for Safe Medication Practices (ISMP). 

Make sure your email is up-to-date
In order to continue to receive updates from MSOS, as well as forum posts and other valuable information as a member of MSOS, please be sure to update your email address with us, whenever it changes. If you need assistance doing so, please send an email to jrufo@ismp.org

4 posts / 0 new
Last post
Gina Gayed
Gina Gayed's picture
Offline
Last seen: 2 days 15 hours ago
Joined: 06/05/2017 - 11:25
NICU 3-mL Doses - Syringe Size and Pump Flow Rate Limitation

Does anyone have any best practices around ensuring 3-mL syringes are used to prepare 3-mL NICU doses?

We’ve had incidents of 3-mL doses dispensed in 5-mL syringes. The ordered rate was less than 0.1 mL/hr. Because the Alaris syringe pump cannot deliver a flow rate lower than 0.1 mL/hr from a 5-mL syringe, the doses had to be sent back to be prepared a second time in 3-mL syringes.

We are thinking about including syringes as components to be scanned when Dispense Prepping NICU mixtures in Epic but are running into some workflow challenges with that.

Appreciate any insight!

Gina Gayed, PharmD, BCPS, CPPS
Medication Safety Officer
Wellstar Health System
Marietta, GA

Tags: